| Posted: Sun Sep 19, 2010 8:14 pm Post subject: |
|
|
OK, I decided I didn't like the motion detection; it didn't seem to be working reliably. So I wrote another bit on Focus!Focus!:
So, y ou can see the webcam window there, the crosshairs showing where I'm recording spots, etc.
It occurs to me that if you pointed the webcam at the tube right, it'd be a good way to make sure your phosphor usage is maximized - a lot easier to tell whether you've got the thing centered by doing that!
You can set how quickly it runs through the graph, choose whether to clear on wrap or not (by not clearing I can leave it running and it'll still show any incidence of dropped contrast). It's kind of specific and doesn't have obvious things like saving, scrolling, etc etc. But for now it works.
It might be interesting to use for doing astig stuff too, since you'd be able to see the history of changes, not just numbers.
Anyway, there you have it. So far no drop outs...
_________________
|
|
| Back to top |
|
 |
perisoft
Joined: 29 Aug 2007 Posts: 2920 Location: Ithaca, NY
|
| Posted: Sun Sep 19, 2010 8:32 pm Post subject: |
|
|
Ah hah! I reproduced the problem while messing with my graphs.
It seems to happen when you change the contrast on the PJ! If I run the contrast up and down,
it'll drop the green out once the contrast hits about 50. In this graph, the first couple are just holding
down the contrast button and running it up as fast as it can go; one is tapping the button so it doesn't
rise as quickly. Rise time appears to affect the likelihood of dropouts, and as you can see, they flicker.
You're seeing the values drop to zero because I'm looking at the bloom from the tube, not the tube
face. It's not dropping to zero.
I checked the tube filaments now that I can trigger the problem - they stay lit. ABG makes no difference
on or off.
This explains why I was seeing it once in a while when testing the projector but not with a static screen.
I tried changing the signal into the PJ but that didn't trigger it - that said, I was turning green off and on,
not ramping it like I was doing for the contrast. That's the next step.
Ideas?
_________________
|
|
| Back to top |
|
 |
CIR Engineering
Joined: 25 Aug 2008 Posts: 4269 Location: Chicago USA & Berlin Germany
|
| Posted: Sun Sep 19, 2010 10:13 pm Post subject: |
|
|
Probably not the tube
Neck card, B board, or a Y board is what I would guess.
craigr
_________________ JETI 1501-HiRes 2nm Spectroradiometer
JETI 1211 Spectroradiometer
Photo Research PR-650 Spectroradiometer
Klein K10-A Colorimeter
Murideo Fresco SIX-G HDMI 2.x Multimedia Generator
Murideo Fresco SIX-A HDMI 2.x Analyzer
Light Illusion ColourSpace XPT Color Calibration Software
Light Illusion LightSpace XPT Pro Version 10.x Color Calibration Software
OMARDRIS JVC Software Patch to use K10-A and Jeti with JVC OEM AutoCal Software!
Sencore CR7000 CRT Tube Analyzer / Rejuvenater
Authorized Dealer for Lumagen & just about everything worth buying
www.CIR-Engineering.com - craigr@cir-engineering.com
Phone: 865-405-6892
|
|
| Back to top |
|
 |
perisoft
Joined: 29 Aug 2007 Posts: 2920 Location: Ithaca, NY
|
| Posted: Sun Sep 19, 2010 11:24 pm Post subject: |
|
|
| CIR Engineering wrote: | Probably not the tube
Neck card, B board, or a Y board is what I would guess.
craigr |
By the way, I'm not sure what you mean by BNC off the B board. There isn't any. Unless you're talking about coax that's terminated differently? But there definitely aren't any BNC connectors around there except on the input side...
Any suggestions as to how I narrow down between B/Y/Neck?
_________________
|
|
| Back to top |
|
 |
CIR Engineering
Joined: 25 Aug 2008 Posts: 4269 Location: Chicago USA & Berlin Germany
|
| Posted: Mon Sep 20, 2010 12:05 am Post subject: |
|
|
Yeah, the minicoax, not BNC.
As I said, swap blue and green neck cards to test the neck cards.
craigr
_________________ JETI 1501-HiRes 2nm Spectroradiometer
JETI 1211 Spectroradiometer
Photo Research PR-650 Spectroradiometer
Klein K10-A Colorimeter
Murideo Fresco SIX-G HDMI 2.x Multimedia Generator
Murideo Fresco SIX-A HDMI 2.x Analyzer
Light Illusion ColourSpace XPT Color Calibration Software
Light Illusion LightSpace XPT Pro Version 10.x Color Calibration Software
OMARDRIS JVC Software Patch to use K10-A and Jeti with JVC OEM AutoCal Software!
Sencore CR7000 CRT Tube Analyzer / Rejuvenater
Authorized Dealer for Lumagen & just about everything worth buying
www.CIR-Engineering.com - craigr@cir-engineering.com
Phone: 865-405-6892
|
|
| Back to top |
|
 |
perisoft
Joined: 29 Aug 2007 Posts: 2920 Location: Ithaca, NY
|
| Posted: Mon Sep 20, 2010 12:27 am Post subject: |
|
|
| CIR Engineering wrote: | Yeah, the minicoax, not BNC.
As I said, swap blue and green neck cards to test the neck cards.
craigr |
10-4.
I wrote a little program to run a variable speed full-screen green ramp-up. It'll reliably trigger the problem, but only if the rate of change is relatively high. If you very gradually increase the APL everything is happy; ramp it quickly and it goes to hell in a handbasket. But it's fine you snap from black to full output suddenly.
Interesting...
_________________
|
|
| Back to top |
|
 |
perisoft
Joined: 29 Aug 2007 Posts: 2920 Location: Ithaca, NY
|
| Posted: Mon Sep 20, 2010 8:29 pm Post subject: |
|
|
OK, I swapped the G and B neck cards. Now the G works fine and the B is *always* dim - very dim; I have to crank its gain all the way up and turn R and G gain down by 70% to get it close to balanced. Ramping up and down contrast on a full white screen doesn't affect any of the channels now - it kind of looks like what was an intermittent problem on that neck card is now a permanent one?
B doesn't seem to be messed up in any -other- way, though. Not flickery or bouncy, etc.
It's possible I didn't get something hooked up right, but I was pretty careful.
Next thing to check for?
_________________
|
|
| Back to top |
|
 |
CIR Engineering
Joined: 25 Aug 2008 Posts: 4269 Location: Chicago USA & Berlin Germany
|
| Posted: Mon Sep 20, 2010 9:48 pm Post subject: |
|
|
So it's almost certainly a problem on the green neck card. I don't have any for the G70, and since this is Curt's forum you should probably see if he can service it for you. Or just see if Curt has one that he can sell outright. The neck boards are pretty much interchangeable between green and blue so he'll probably have one he can sell... and they are not too expensive either.
If you are at all good with soldering, you may want to look over all the solder joins on the green neck card before you write it off. There very likely could just be a bad solder joint, and a quick touch with the iron may fix it good as new.
Honestly, this is about the very best outcome you could have hoped for An easy to find board that is not expensive.
craigr
_________________ JETI 1501-HiRes 2nm Spectroradiometer
JETI 1211 Spectroradiometer
Photo Research PR-650 Spectroradiometer
Klein K10-A Colorimeter
Murideo Fresco SIX-G HDMI 2.x Multimedia Generator
Murideo Fresco SIX-A HDMI 2.x Analyzer
Light Illusion ColourSpace XPT Color Calibration Software
Light Illusion LightSpace XPT Pro Version 10.x Color Calibration Software
OMARDRIS JVC Software Patch to use K10-A and Jeti with JVC OEM AutoCal Software!
Sencore CR7000 CRT Tube Analyzer / Rejuvenater
Authorized Dealer for Lumagen & just about everything worth buying
www.CIR-Engineering.com - craigr@cir-engineering.com
Phone: 865-405-6892
|
|
| Back to top |
|
 |
perisoft
Joined: 29 Aug 2007 Posts: 2920 Location: Ithaca, NY
|
| Posted: Mon Sep 20, 2010 11:07 pm Post subject: |
|
|
| CIR Engineering wrote: | | There very likely could just be a bad solder joint, and a quick touch with the iron may fix it good as new. |
Presumably there are fairly few places in the circuit that could cause dim but correct output. Guess I'll have to take a look around.
| Quote: |
Honestly, this is about the very best outcome you could have hoped for An easy to find board that is not expensive.
craigr |
Well, the best outcome would have been it being a loose wire, but aside from that... The fact that the nature of the problem changed when I swapped boards suggests something along those lines.
I also know more about the inside of a G70 - which, by the way, is really nice to work in; that's something I noticed with the 1271 too. The wiring is really neat, harnessed well, the connectors are well thought out... Why, it's almost like they knew you'd be spending a lot of time in there!
One other thing - I've got to check to see if the board is seated well enough on the tube. It's a bit further out than the G and R. If it becomes intermittent again instead of always low, that might suggest a pin contact problem.
Another quick question - you said that the B and G neck cards are 'pretty much the same' - so they're different than the red significantly? Different output levels expected, or something?
_________________
|
|
| Back to top |
|
 |
CIR Engineering
Joined: 25 Aug 2008 Posts: 4269 Location: Chicago USA & Berlin Germany
|
| Posted: Tue Sep 21, 2010 1:11 pm Post subject: |
|
|
| perisoft wrote: |
Another quick question - you said that the B and G neck cards are 'pretty much the same' - so they're different than the red significantly? Different output levels expected, or something? |
On the G90 the red tube is oriented 180 degrees upside down compared to the blue and green. This means that the neck card on the red has all its plugs on the opposite side compared to the blue and green.
I think this is the case on the G70 as well, but I am not positive so you can check for yourself. Everything else is pretty much the same, so you could even use a red neck card if you can find enough slack in the wires to make the connections. In the G90 this isn't usually possible without cutting and pasting wires.
See the G90 also uses a very small actual neck card. The real guts of the neck cards are located on two additional boards that are above the neck cards. The K and G1 then is delivered to the small neck card fully amplified so all the small neck card does is connect to the tube. There really isn't anything to fail on this tiny neck card so they never get swapped because they never die.
The G70 is more traditional with the video amps and other hardware on the actual neck card so maybe it's not the same as the G90 in terms of swapping and maybe they have the wiring plugs in the same locations? I can't remember.
craigr
_________________ JETI 1501-HiRes 2nm Spectroradiometer
JETI 1211 Spectroradiometer
Photo Research PR-650 Spectroradiometer
Klein K10-A Colorimeter
Murideo Fresco SIX-G HDMI 2.x Multimedia Generator
Murideo Fresco SIX-A HDMI 2.x Analyzer
Light Illusion ColourSpace XPT Color Calibration Software
Light Illusion LightSpace XPT Pro Version 10.x Color Calibration Software
OMARDRIS JVC Software Patch to use K10-A and Jeti with JVC OEM AutoCal Software!
Sencore CR7000 CRT Tube Analyzer / Rejuvenater
Authorized Dealer for Lumagen & just about everything worth buying
www.CIR-Engineering.com - craigr@cir-engineering.com
Phone: 865-405-6892
|
|
| Back to top |
|
 |
perisoft
Joined: 29 Aug 2007 Posts: 2920 Location: Ithaca, NY
|
| Posted: Fri Sep 24, 2010 2:07 am Post subject: |
|
|
The neck cards all look the same. The R and G are marked (by hand) though; either there's a small difference or someone was cautious.
I swapped G and B; and like I said, the problem went to B but is no longer intermittent.
I decided to swap back to make sure there's not something else wacky going on; the problem moved to the green tube very nicely; still constant. It seems a bit worse than it did with B, but presumably they respond in somewhat different ways...
So yeah, pretty much I gotta get a neck card, I guess. Or get this one repaired. So, until I can find the time to get all that done, the Sony waits... Hopefully a fix will get it working well.
_________________
|
|
| Back to top |
|
 |
CIR Engineering
Joined: 25 Aug 2008 Posts: 4269 Location: Chicago USA & Berlin Germany
|
| Posted: Fri Sep 24, 2010 5:06 am Post subject: |
|
|
Neck card is no big deal. Check for a bad solder joint first, or just buy a new one from Curt.
craigr
_________________ JETI 1501-HiRes 2nm Spectroradiometer
JETI 1211 Spectroradiometer
Photo Research PR-650 Spectroradiometer
Klein K10-A Colorimeter
Murideo Fresco SIX-G HDMI 2.x Multimedia Generator
Murideo Fresco SIX-A HDMI 2.x Analyzer
Light Illusion ColourSpace XPT Color Calibration Software
Light Illusion LightSpace XPT Pro Version 10.x Color Calibration Software
OMARDRIS JVC Software Patch to use K10-A and Jeti with JVC OEM AutoCal Software!
Sencore CR7000 CRT Tube Analyzer / Rejuvenater
Authorized Dealer for Lumagen & just about everything worth buying
www.CIR-Engineering.com - craigr@cir-engineering.com
Phone: 865-405-6892
|
|
| Back to top |
|
 |
|
|
|
|
|
You cannot post new topics in this forum You cannot reply to topics in this forum You cannot edit your posts in this forum You cannot delete your posts in this forum You cannot vote in polls in this forum You cannot attach files in this forum You can download files in this forum
|
Forum powered by phpBB © phpBB Group
|
|