| HK-Steve wrote: | Interesting that all of these HVPS, none have any faulty potted sections, I think,
No smoke, or breakdown of the potted section. |
If the fault is within the HV transformer (shortened windings) and/or multiplier section (shortened diode) You might not even be able to see it after having opened the potted section. Often the defect only will become apparent performing adecuate measurements.
Anyway: If the fuse blows the this is a clear indicator of an overload event or a short. Have You measured the resistance of the MOSFETs between D and S (Drain & Source)? You could compare Your readings with those of a known intact HVPS...
Regards,
barclay66
|
|
| Back to top |
|
 |
barclay66
Joined: 27 Jun 2011 Posts: 1304 Location: Germany
TV/Projector: Marquee 9500 Ultra
|
| Posted: Tue Jan 10, 2012 8:28 pm Post subject: |
|
|
Uh, lots of posts to answer...
| HK-Steve wrote: | BUT, it seems more like good luck, than management if the sockets are removed, ie, no pattern.
34Kv, Rev. C4, sockets
34.9Kv Rev.A1, NO sockets
34.9Kv Rev.A7, sockets
34.9Kv Rev.A7 NO sockets
34.9Kv Rev.B series, all NO sockets
|
I do see a pattern here:
In general, the newer the less probable sockets became.
At a specific point the new PCB without sockets started production. Older PCBs had to processed anyway, just that there the diodes were left unpopulated. Maybe at the time Rev.A7 was produced someone found old stock of older PCBs or there was a shortage of the newer ones. Nothing unusual in series production...
Regards,
barclay66
|
|
| Back to top |
|
 |
HK-Steve
Joined: 15 Jul 2006 Posts: 849 Location: Switzerland
TV/Projector: Marquee 9500, Epson 8100
|
| Posted: Wed Jan 11, 2012 2:27 pm Post subject: |
|
|
Michael, sorry to say it, but,
Pop goes the HVPS, again, and again, and again.
replaced the fuse, powered up with no tubes connected, splitter only, fuse blew straight away.
Same symptoms as the last 2 HVPS's.
Will see if I have a working A Revision.
Cheers
_________________
|
|
| Back to top |
|
 |
Nashou66
Joined: 12 Jan 2007 Posts: 16171 Location: West Seneca NY
|
| Posted: Wed Jan 11, 2012 2:51 pm Post subject: |
|
|
Are you sure the 390 volt line is not compromised? Do you have a spare Wiring harness you can replace Michael's with? It just seems wrong that its blowing all the time. Or maybe the back plane has a problem in the traces or the connectors. I would just do a tear down
to check it out. I hate seeing HVPS being eaten up by this machine.
Maybe replace his chassis with another one from an 8500.
Athanasios
_________________ Don't blame your underwear for your crooked ass~ unknown Greek philosopher
"Republicans believe every day is the Fourth of July, but the Democrats believe every day is April 15." --- President Reagan
One Smart Dog!!!
Marquee High Performance Bellows now shipping!!
Marquee Modifications and Performance Enhancement
Marquee C-element and Bellow removal
|
|
| Back to top |
|
 |
HK-Steve
Joined: 15 Jul 2006 Posts: 849 Location: Switzerland
TV/Projector: Marquee 9500, Epson 8100
|
| Posted: Wed Jan 11, 2012 2:58 pm Post subject: |
|
|
Thanks for the suggestions Athanasios, but,
I had Michaels chassis at my place, replaced the loom, and also checked the LVPS output,
It sat stable for 8hrs at 383 to 385volts, as I data logged it when running for a full day.
Back plane board was also checked, but looked perfect.
I do not have any other chassis to change-over, don't see how it would help??
Appreciate your comments, but looks like I will loose even more hair with this problem.
Cheers
_________________
|
|
| Back to top |
|
 |
Nashou66
Joined: 12 Jan 2007 Posts: 16171 Location: West Seneca NY
|
| Posted: Wed Jan 11, 2012 3:10 pm Post subject: |
|
|
Steve when checking the 390 volt line are you checking it where it comes out of the LVPS location or on the output socket to the HVPS?
Id check it at the HVPS end.
This is f*ck*** up, I read back and see it worked with the 34kv PS but it got too hot to touch and you had to shut it down.
I wonder if there is anything under the mother board or behind the back plane that you cant see causing a short.
If it was me id tear down the entire chassis and double check everything with a fine tooth comb .
The reason id try another chassis is because there has to be a short somewhere with the 390vdc line/trace or one of the G2 lines/trace.
that chassis has bad JuJu now!!!
EDIT: I had a set that kept showing a fail led on the 24 volt line. i couldn't find the issue. so i tore the set apart to part it out and there was a sheet metal screw behind the Back plane. I put it all back together and the LVPS had no error lights.
Athanasios
_________________ Don't blame your underwear for your crooked ass~ unknown Greek philosopher
"Republicans believe every day is the Fourth of July, but the Democrats believe every day is April 15." --- President Reagan
One Smart Dog!!!
Marquee High Performance Bellows now shipping!!
Marquee Modifications and Performance Enhancement
Marquee C-element and Bellow removal
|
|
| Back to top |
|
 |
Curt Palme CRT Tech
Joined: 08 Mar 2006 Posts: 24396 Location: Langley, BC
TV/Projector: All of them!
|
| Posted: Wed Jan 11, 2012 3:17 pm Post subject: |
|
|
I haven't joined in here, because I have nothing really to add.
Other than the 390 volts input, there's nothing here really on the backplane or control circuits that affect the HVPS other than turning it on and off.
Here's the thing though about the shorted MOSFETs: It takes a LOT usually to create shorted windings in a transformer, as in the one that couples the output of the MOSFETs to the potted HV section. It's a coupling transformer. I know I had a bunch in the older Ampro 2000 HVPS, and have indeed seen some in the Marquee supplies, but it's sure not a consistent failure.
As I said above, I've had one Marquee do the same thing, and it never got resolved. Unless Scott has something to add here that they may have seen at the factory, I got nothing, sorry!
|
|
| Back to top |
|
 |
HK-Steve
Joined: 15 Jul 2006 Posts: 849 Location: Switzerland
TV/Projector: Marquee 9500, Epson 8100
|
| Posted: Wed Jan 11, 2012 3:23 pm Post subject: |
|
|
Appreciate any help Curt, and you have helped, like everyone else who has replied.
So what did you do with the Marquee that had same kind of HV problem?
Cheers
_________________
|
|
| Back to top |
|
 |
Curt Palme CRT Tech
Joined: 08 Mar 2006 Posts: 24396 Location: Langley, BC
TV/Projector: All of them!
|
| Posted: Wed Jan 11, 2012 3:36 pm Post subject: |
|
|
I don't know what the customer did with it. It was a HVPS sale only, 30 days later he told me it failed again, I sent a second replacement. 30 days later he called again. I asked how many he'd gone through, and he said 5 or 6. IN 6 months. (!!) Yeah, no warranty, sorry!
He did send me one of my sold ones back along with one of his other ones, but he'd hacked his potted section apart and damaged a bunch of the parts, so I had no idea what the problem was. I believe one of mine that he'd sent back had the 'ticking' sound in the potted section, so I threw it out.
WIth shorted MOSFETs, I've had about 50% success in replacing just the MOSFETs again. 50% of the time, they run red hot, indicating a shorted output transformer. I used to scope the gates of the MOSFETs to see if the input waveform had changed, but I think I've only ever had one bad HVPS driver board, and that caused a dead HVPS.
|
|
| Back to top |
|
 |
flipdog
Joined: 22 Dec 2011 Posts: 156
|
| Posted: Wed Jan 11, 2012 4:49 pm Post subject: |
|
|
|
Possible arc back through the G2 line.........
|
|
| Back to top |
|
 |
HK-Steve
Joined: 15 Jul 2006 Posts: 849 Location: Switzerland
TV/Projector: Marquee 9500, Epson 8100
|
| Posted: Thu Jan 12, 2012 10:15 am Post subject: |
|
|
Would an arc through the G2, damage the VNB?
Next step, is to work out which tube,
Tubes were test using the Sencore which checked out all good.
Don't know what to do now.
Cheers
_________________
|
|
| Back to top |
|
 |
HK-Steve
Joined: 15 Jul 2006 Posts: 849 Location: Switzerland
TV/Projector: Marquee 9500, Epson 8100
|
| Posted: Thu Jan 12, 2012 10:32 am Post subject: |
|
|
Athanasios,
I might get the projector down and back to my place,
but I pretty much checked everything the first time, as per your suggestion,
might be the only choice to finally work out the problem.
Cheers
_________________
|
|
| Back to top |
|
 |
Curt Palme CRT Tech
Joined: 08 Mar 2006 Posts: 24396 Location: Langley, BC
TV/Projector: All of them!
|
| Posted: Thu Jan 12, 2012 3:33 pm Post subject: |
|
|
| flipdog wrote: | | Possible arc back through the G2 line......... |
Hey, anything is possible, but remember that the G2 lines are isolated by the G2 controls, transistors, resistors, etc. I believe that even a dead short on the G2 output won't do anything, but as I said, anything is possible. I'd be curious to see if the HV lead from the HVPS plugged into the splitter, leaving the G2 lines disconnected and the tube HV leads out of the splitter would cause an HVPS failure?
|
|
| Back to top |
|
 |
Nashou66
Joined: 12 Jan 2007 Posts: 16171 Location: West Seneca NY
|
|
| Back to top |
|
 |
barclay66
Joined: 27 Jun 2011 Posts: 1304 Location: Germany
TV/Projector: Marquee 9500 Ultra
|
| Posted: Thu Jan 12, 2012 4:28 pm Post subject: |
|
|
| Nashou66 wrote: | | I still say its the 390 volt line, or the back plane connector for the HVPS. Something is causing a short to the HVPS form the outside. |
Hi,
How could a short on the 390V line outside make the fuse inside the HVPS blow? If the supply line gets shorted before going into the HVPS then it wouldn't even reach the fuse inside. Such a short would more likely make the LVPS shut down (protection) or damage it.
A backplane connector problem seems much more likely. What if there was a weak contact which intermittently would open? If this happened on the 390V line the following could happen:
- Contact resistance increases
- Supply voltage drops
- HV regulation tries to compensate (increase of output power)
- HV regulation goes to full throttle
- Contact gets hot due to higher load
- A small arc on the contact occurs
- The arc re-establishes the contact (oxidized surface burnt away)
- Supply voltage back to normal
- HV regulation still on full throttle
- *POP*
I would closely inspect the connector surfaces on the dead HVPS and even more the backplane connector...
Regards,
barclay66
|
|
| Back to top |
|
 |
Nashou66
Joined: 12 Jan 2007 Posts: 16171 Location: West Seneca NY
|
| Posted: Thu Jan 12, 2012 4:59 pm Post subject: |
|
|
What i meant it is not caused inside the HVPS, all the HV supplies would work in another chassis i bet and would not blow as would their LVPS's. So what I meant was what you're saying is the back plane connector at the 390 volt output must be shorting the HVPS. I couldn't see it being the G2 connections on the Back plane. It blows the fuse to fast, and I assume not giving the G2 time to cause the failure/Short.
Athanasios
_________________ Don't blame your underwear for your crooked ass~ unknown Greek philosopher
"Republicans believe every day is the Fourth of July, but the Democrats believe every day is April 15." --- President Reagan
One Smart Dog!!!
Marquee High Performance Bellows now shipping!!
Marquee Modifications and Performance Enhancement
Marquee C-element and Bellow removal
|
|
| Back to top |
|
 |
barclay66
Joined: 27 Jun 2011 Posts: 1304 Location: Germany
TV/Projector: Marquee 9500 Ultra
|
| Posted: Fri Jan 13, 2012 9:12 am Post subject: |
|
|
| HK-Steve wrote: | Is it possible to adjust the 34Kv to 34.9Kv, with the right test gear of course !!!!!!!!
There are 3 trim pots on the HVPS, which does what?
or is it in the potted section that holds the secret of the output voltage??? |
Hi Steve,
I've sent a PM to You regarding this topic...
Regards,
barclay66
|
|
| Back to top |
|
 |
HK-Steve
Joined: 15 Jul 2006 Posts: 849 Location: Switzerland
TV/Projector: Marquee 9500, Epson 8100
|
| Posted: Fri Jan 13, 2012 10:29 am Post subject: |
|
|
PM, right back at you.
Chers
_________________
|
|
| Back to top |
|
 |
HK-Steve
Joined: 15 Jul 2006 Posts: 849 Location: Switzerland
TV/Projector: Marquee 9500, Epson 8100
|
| Posted: Fri Jan 13, 2012 10:41 am Post subject: |
|
|
The contacts on all 3 HVPS's are perfect,
shiny, no burns, no tarnish, nothing, see pic, just a nice pressure contact mark.
all look exactly the same.
Keep the ideas comming, we are going to get this worked out.
Cheers
_________________
|
|
| Back to top |
|
 |
barclay66
Joined: 27 Jun 2011 Posts: 1304 Location: Germany
TV/Projector: Marquee 9500 Ultra
|
| Posted: Fri Jan 13, 2012 2:30 pm Post subject: |
|
|
Hi,
The backplane connector is part of the wiring harness which is located below the belly fans. Although the connector itself seems to be ok it could be possible that one of the wires hasn't been crimped to it correctly. This could be checked. Another idea could be changing the complete harness (maybe some can take one out from a scrapped set)...
Regards,
barclay66
|
|
| Back to top |
|
 |
|
|
|
|
|